A SERVICE OF

logo

Getting Started Teledyne API - T100 UV Fluorescence SO2 Analyzer
34
Verify that there is no apparent external shipping damage. If damage has occurred,
please advise the shipper first, then Teledyne API.
Included with your analyzer is a printed record of the final performance characterization
performed on your instrument at the factory. It is titled Final Test and Validation Data
Sheet (P/N 04551). This record is an important quality assurance and calibration record
for this instrument. It should be placed in the quality records file for this instrument.
With no power to the unit, carefully remove the top cover of the analyzer and check for
internal shipping damage by carrying out the following steps:
1. Remove the locking screw located in the top, center of the Front panel;
2. Remove the two flat head, Phillips screws on the sides of the instrument;
3. Slide the cover backwards until it clears the analyzer’s front bezel, and;
4. Lift the cover straight up.
5. Inspect the interior of the instrument to ensure that all circuit boards and other
components are in good shape and properly seated.
6. Check the connectors of the various internal wiring harnesses and pneumatic
hoses to ensure that they are firmly and properly seated.
7. Verify that all of the optional hardware ordered with the unit has been installed.
These are listed on the paperwork accompanying the analyzer.
WARNING
ELECTRICAL SHOCK HAZARD
Never disconnect PCAs, wiring harnesses or electronic subassemblies
while under power.
3.1.1. VENTILATION CLEARANCE
Whether the analyzer is set up on a bench or installed into an instrument rack, be sure to
leave sufficient ventilation clearance.
Table 3-1: Ventilation Clearance
AREA
MINIMUM REQUIRED
CLEARANCE
Back of the instrument
4 in.
Sides of the instrument
1 in.
Above and below the
instrument
1 in.
Various rack mount kits are available for this analyzer. Refer to Section 1.4 of this
manual for more information.