A SERVICE OF

logo

APPENDIX A - Version Specific Software Documentation
Rev 1.0.3 (T-Series)/G6 (E-Series)
Teledyne API - Models T100, 100E Series (05036F DCN6650)
A-14 05036 Rev D
SETUP VARIABLE
NUMERIC
UNITS
DEFAULT
VALUE
VALUE RANGE DESCRIPTION
SLOPE_CONST — 8,
6.25
3
0.1–10
Constant to make visible slope close to
1.
DARK_ENABLE — ON,
OFF
12
OFF, ON
ON enables PMT/UV dark calibration;
OFF disables it.
DARK_FREQ Minutes 30,
720
3
0.1–1440 Dark calibration period.
DARK_PRE_DWELL Seconds 10 1–60
Dwell time after closing dark shutter or
turning off lamp or selecting preamp
range.
DARK_POST_DWELL Seconds 10,
30
3
1–180
Dwell time after opening dark shutter or
turning on lamp.
DARK_SAMPLES Samples 5 1–10 Number of dark samples to average.
DARK_FSIZE Samples 2 1–100 Dark offset moving average filter size.
DARK_LIMIT mV 200,
400
3
0–1000 Maximum dark offset allowed.
SO2_TARG_ZERO1 Conc 0 -100–999.99
Target SO
2
concentration during zero
calibration of range 1.
SO2_SPAN1 Conc 400,
4000
3
0.01–9999.99
Target SO
2
concentration during span
calibration of range 1.
SO2_SLOPE1 PPB/mV,
PPM/mV
3
1 0.25–4 SO
2
slope for range 1.
SO2_OFFSET1 mV 0 -55000–55000 SO
2
offset for range 1.
SO2_BACKGROUND1
4
PPM 0 -10000–10000
SO
2
background concentration for
range 1.
SO2_TARG_ZERO2 Conc 0 -100–999.99
Target SO
2
concentration during zero
calibration of range 2.
SO2_SPAN2 Conc 400,
4000
3
0.01–9999.99
Target SO
2
concentration during span
calibration of range 2.
SO2_SLOPE2 PPB/mV,
PPM/mV
3
1 0.25–4 SO
2
slope for range 2.
SO2_OFFSET2 mV 0 -55000–55000 SO
2
offset for range 2.
SO2_BACKGROUND2
4
PPM 0 -10000–10000
SO
2
background concentration for
range 2.
CO2_TARG_SPAN_CONC
11
% 12 0.1–1000
Target CO
2
concentration during span
calibration.
CO2_SLOPE
11
1 0.5–5 CO
2
slope.
CO2_OFFSET
11
% 0 -10–10 CO
2
offset.
O2_TARG_SPAN_CONC
10
% 20.95 0.1–100
Target O
2
concentration during span
calibration.
O2_SLOPE
10
1 0.5–2 O
2
slope.
O2_OFFSET
10
% 0 -10–10 O
2
offset.
RANGE_MODE — SNGL SNGL,
DUAL,
AUTO
Range control mode. Enclose value in
double quotes (“) when setting from the
RS-232 interface.
06807C DCN6650