APPENDIX A - Version Specific Software Documentation
Rev 1.0.3 (T-Series)/G6 (E-Series)
Teledyne API - Models T100, 100E Series (05036F DCN6650)
A-18 05036 Rev D
SETUP VARIABLE
NUMERIC
UNITS
DEFAULT
VALUE
VALUE RANGE DESCRIPTION
RCELL_INTEG — 0.005 0–10
Reaction cell temperature PID integral
coefficient.
RCELL_DERIV — 0.5 0–10
Reaction cell temperature PID
derivative coefficient.
O2_CELL_CYCLE
10
Seconds 10 0.5–30
O
2
cell temperature control cycle
period.
O2_CELL_PROP
10
— 1 0–10
O
2
cell PID temperature control
proportional coefficient.
O2_CELL_INTEG
10
— 0.1 0–10
O
2
cell PID temperature control integral
coefficient.
O2_CELL_DERIV
10
— 0 (disabled) 0–10
O
2
cell PID temperature control
derivative coefficient.
IZS_CYCLE
1
Seconds 2 0.5–30 IZS temperature control cycle period.
IZS_PROP
1
1/ºC
1 (prop. band
= 1 ºC)
0–10
IZS temperature PID proportional
coefficient.
IZS_INTEG
1
— 0.03 0–10
IZS temperature PID integral
coefficient.
IZS_DERIV
1
— 0 0–10
IZS temperature PID derivative
coefficient.
650,
550
3
HVPS_SET Volts
Warnings:
400–900,
400–700
3
0–2000
High voltage power supply warning
limits. Set point is not used.
MAX_PMT_DETECTOR mV 4995 0–5000 PMT detector maximum warning limit.
450 PHOTO_ABS_LIMITS
1
mV
Warnings:
125–625
0–5000
Pre-amplified UV lamp
minimum/maximum warning limits. Set
point is not used.
3500 UV_LAMP_LIMITS mV
Warnings:
1000–4995
0–5000
UV lamp minimum/maximum warning
limits. Set point is not used.
ELEC_TEST_LEVEL
12
— 0 0–65535 Electrical test level setting.
OPTIC_TEST_LEVEL
12
— 0 0–65535 Optical test level setting.
CONC_LIN_ENABLE
3
— ON OFF, ON
ON enables concentration linearization;
OFF disables it.
STAT_REP_PERIOD
17
Seconds 1 0.5–120
TAI protocol status message report
period.
SERIAL_NUMBER — “00000000 ”
Any character in
the allowed
character set. Up to
100 characters
long.
Unique serial number for instrument.
Enclose value in double quotes (“)
when setting from the RS-232
interface.
DISP_INTENSITY — HIGH HIGH,
MED,
LOW,
DIM
Front panel display intensity. Enclose
value in double quotes (“) when setting
from the RS-232 interface.
I2C_RESET_ENABLE — ON OFF, ON I
2
C bus automatic reset enable.