A SERVICE OF

logo

Calibration Procedures Teledyne API - T100 UV Fluorescence SO2 Analyzer
198
9.7. MANUAL CALIBRATION IN DUAL OR AUTO REPORTING
RANGE MODES
When the analyzer is in either Dual or Auto Range modes the user must run a separate
calibration procedure for each range. After pressing the CAL, CALZ or CALS buttons
the user is prompted for the range that is to be calibrated as seen in the CALZ example
below:
Continue Calibration as per
Standard Procedure
SAMPLE RANGE = 500.000 PPB SO2 =XXX.X
< TST TST > CAL CALZ CALS SETUP
ZERO CAL M RANGE = 500.000 PPB SO2 =XXX.X
< TST TST > ZERO CONC EXIT
WAIT 10
MINUTES
Or until the
reading
stabilizes and
the ZERO button
is displayed
SAMPLE RANGE TO CAL: LOW
LOW HIGH ENTR SETUP
SAMPLE RANGE TO CAL: HIGH
LOW HIGH ENTR SETUP
Analyzer enters
ZERO CAL Mode
Refer to Table 6-1 for
Z/S Valve States
during this operating
mode.
Figure 9-11: Manual Calibration in Dual/Auto Reporting Range Modes
Once this selection is made, the calibration procedure continues as previously described
in Sections 7.2 through 7.6. The other range may be calibrated by starting over from the
main SAMPLE display.
9.7.1. CALIBRATION WITH REMOTE CONTACT CLOSURES
Contact closures for controlling calibration and calibration checks are located on the rear
panel CONTROL IN connector. Instructions for setup and use of these contacts can be
found in Section 8.1.2.
When the appropriate contacts are closed for at least 5 seconds, the instrument switches
into zero, low span or high span m
ode and the internal zero/span valves will be
automatically switched to the appropriate configuration. The remote calibration contact
closures may be activated in any order. It is recommended that contact closures remain
closed for at least 10 minutes to establish a reliable reading; the instrument will stay in
the selected mode for as long as the contacts remain closed.
06807C DCN6650