5.9.7. PRESSURE CALIBRATION
A sensor at the exit of the sample chamber continuously measures the pressure of the
sample gas. This data is used to compensate the final SO
2
concentration calculation for
changes in atmospheric pressure when the instrument’s TPC feature is turned on (refer
to Section 10.7.3) and is stored in the CPU’s memory as the test function PRES (also
viewable via the front panel).
Ensure to use a barometer that measures actual barometric pressure.
To cause the analyzer to measure and record a value for PRES, access the Signal I/O
from the DIAG Menu (refer to Figure 5-16), then press:
DIAG SIGNAL I / O
NEXT ENTR EXIT
Repeat Pressing NEXT until . .
.
DIAG PRESSURE CALIBRATION
PREV NEXT ENTR EXIT
DIAG PCAL ACTUAL PRES :27.20 IN-HG-A
2 7 .2 0 ENTR EXIT
ENTR accepts the
new value
EXIT ignores the new
value
Adjust these values until the
displayed pressure equals the
pressure measured by the
independent pressure meter.