![](https://pdfstore-manualsonline.prod.a.ki/pdfasset/7/7f/77f9fc71-f015-46f6-b031-e970dc9e3f85/77f9fc71-f015-46f6-b031-e970dc9e3f85-bg7d.png)
Teledyne API - T100 UV Fluorescence SO2 Analyzer SETUP Menu
123
5.9.3.6. AIN CALIBRATION
This is the sub-menu to conduct the analog input calibration. This calibration should
only be necessary after major repair such as a replacement of CPU, motherboard or
power supplies. Navigate to the ANALOG I/O CONFIGURATION MENU from the
DIAG Menu (refer to Figure 5-16), then press:
Exit at any time to
return to the main
DIAG menu
DIAG
NALOG I / O CONFIGURATION
PREV NEXT ENTR EXIT
Continue pressing SET> until …
DIAG AIO
IN CALIBRATED: NO
< SET SET> CAL EXIT
DIAG AIO AIN CALIBRATED: YES
< SET SET> CAL EXIT
Instrument
calibrates
automatically
Exit to return to the
ANALOG I/O
CONFIGURATION
MENU
DIAG AIO CALIBRATING A/D ZERO
CALIBRATING A/D SPAN
Figure 5-28: DIAG – Analog Output – AIN Calibration
5.9.3.7. ANALOG INPUTS (XIN1…XIN8) OPTION CONFIGURATION
To configure the analyzer’ optional analog inputs define for each channel:
gain (number of units represented by 1 volt)
offset (volts)
engineering units to be represented in volts (each press of the touchscreen button
scrolls the list of alphanumeric characters from A-Z and 0-9)
whether to display the channel in the Test functions