Section II Operating Instructions Teledyne API T802 Paramagnetic O
2
Analyzer Operation Manual
58
3.4.3.2. DILUTION RATIO (OPTION) SET UP
If the dilution ratio option is enabled on your T802 and your application involves diluting the sample gas
before it enters the analyzer, set the dilution ratio as follows:
EXAMPLE
SAMPLE RANGE=100.00 % O2=XXX.XX
< TST TST > CAL SETUP
SETUP X.X PRIMARY SETUP MENU
CFG ACAL DAS RNGE PASS CLK MORE EXIT
SETUP X.X RANGE CONTROL MENU
MODE SET DIL EXIT
SETUP X.X O2 DIL FACTOR:1.0 Gain
0 0 0 0 1 .0 ENTR EXIT
SETUP X.X O2 DIL FACTOR 10.0 Gain
0 0 0 1 0 .0 ENTR EXIT
Toggle these buttons to
set the dilution factor.
This is the number by
which the analyzer will
multiply the O
2
concentration of the gas
passing through the
reaction cell.
EXIT ignores the
new setting.
ENTR accepts the
new setting.
SETUP X.X CO2 DIL FACTOR:1.0 Gain
0 0 0 0 1 .0 ENTR EXIT
Only appears if the
optional CO
2
sensor
is installed.
07275B DCN6418