A SERVICE OF

logo

07275A (DCN6005) 185
10. MAINTENANCE SCHEDULE & PROCEDURES
The T802 Paramagnetic Oxygen Analyzer utilizes a technology that is non-
depleting and requires very little maintenance. However, there are a minimal
number of simple procedures that when performed regularly will ensure that the
analyzer continues to operate accurately and reliably over its lifetime. Repairs
and troubleshooting are covered in Section 11 of this manual.
10.1. MAINTENANCE SCHEDULE
Table 10-1 shows a typical maintenance schedule for the analyzer. Please note
that in certain environments (i.e. dusty, very high ambient pollutant levels) some
maintenance procedures may need to be performed more often than shown.
Note
A Span and Zero Calibration Check (see CAL CHECK REQ’D Column of Table
10-1) must be performed following certain of the maintenance procedure
listed below.
HAZARD
STRONG OXIDIZER
OXYGEN IS A STRONG OXIDIZER.
Before working with the casing open, be sure to turn off power supply, and perform
air or N
2
gas purging of not only the analyzer inside, but also the sample gas line.
In addition, carefully prevent oil and grease from adhering to any piping. Otherwise,
poisoning, fire or explosion may be caused due to gas leakage, etc.
CAUTION
GENERAL SAFETY HAZARD
Risk of electrical shock. Disconnect power before performing any of the following operations
that require entry into the interior of the analyzer.
CAUTION
QUALIFIED PERSONNEL
The operations outlined in this Section are to be performed by qualified maintenance
personnel only.
07275B DCN6418