Teledyne API T802 Paramagnetic O
2
Analyzer Operation Manual
39
Figure 3-11: Concentration Alarm Relay
Alarm 1 “System OK 2”
Alarm 2 “Conc 1”
Alarm 3 “Conc 2”
Alarm 4 “Range Bit”
“ALARM 1” RELAY
Alarm 1 which is “System OK 2” (system OK 1, is the status bit) is in the
energized state when the instrument is “OK” & there are no warnings. If there is
a warning active or if the instrument is put into the “DIAG” mode, Alarm 1 will
change states. This alarm has “reverse logic” meaning that if you put a meter
across the Common & Normally Closed pins on the connector you will find that
it is OPEN when the instrument is OK. This is so that if the instrument should
turn off or lose power, it will change states & you can record this with a data
logger or other recording device.
“ALARM 2” RELAY & “ALARM 3” RELAY
The “Alarm 2 Relay” on the rear panel, is associated with the “Concentration
Alarm 1” set point in the software & the “Alarm 3 Relay” on the rear panel is
associated with the “Concentration Alarm 2” set point in the software.
Alarm 2 Relay O
2
Alarm 1 = xxx %
Alarm 3 Relay O
2
Alarm 2 = xxx %
Alarm 2 Relay CO
2
Alarm 1 = xxx % (If CO
2
option is present)
Alarm 3 Relay CO
2
Alarm 2 = xxx % (If CO
2
option is present)
The Alarm 2 Relay will be turned on any time the concentration set-point is
exceeded & will return to its normal state when the concentration value goes back
below the concentration set-point.
Even though the relay on the rear panel is a NON-Latching alarm & resets when
the concentration goes back below the alarm set point, the warning on the front
panel of the instrument will remain latched until it is cleared. You can clear the
warning on the front panel by either pushing the CLR button on the front panel or
through the serial port.
The software for this instrument is flexible enough to allow you to configure the
alarms so that you can have 2 alarm levels for each gas.
O
2
Alarm 1 = 20 %
O
2
Alarm 2 = 100 %
CO
2
Alarm 1 = 20 %
CO
2
Alarm 2 = 100 %