Teledyne API T802 Paramagnetic O
2
Analyzer Operation Manual
53
3.4.1.1. WARNING MESSAGES
Because internal temperatures and other conditions may be outside the specified
limits during the analyzer’s warm-up period, the software will suppress most
warning conditions for 30 minutes after power up. If warning messages persist
after the 60-minute warm up period, investigate their cause using the
troubleshooting guidelines in Section 11.
To view and
clear w
arning messages, press:
Suppresses the
warning messages
Press CLR to clear the current
message.
If more than one warning is
active, the next message will take
its place.
Once the last warning has
been cleared, the RANGE
function will be displayed in
the analyzer’s main
MESSAGE FIELD.
NOTE:
If a warning message persists after
several attempts to clear it, the message
may indicate a real problem and not an
artifact of the warm-up period
SAMPLE SYSTEM RESET
TEST CAL MSG CLR SETUP
SAMPLE SYSTEM RESET
TEST CAL MSG CLR SETUP
SAMPLE SYSTEM RESET
TEST CAL MSG CLR SETUP
SYSTEM SYSTEM RESET
TEST CLR SETUP
MSG returns the active
warnings to the message
field.
STANDBY RANGE=100.00 % O2=XXX.XX
TEST CAL MSG SETUP
Figure 3-19: Viewing and Clearing T802 WARNING Messages
Table 3-10 lists brief descriptions of the warning messages that may occur during
startup.