A SERVICE OF

logo

Models T802, 802E Appendix A Menu Trees (Reference: 06530C DCN6418) Appendix A
Error! Unknown document property name.Error! Unknown document property name. A-17
APPENDIX A-3: Warnings and Test Measurements, Software Versions 1.0.3 (T-Series)/A.3 (E-Series)
Table A-2: Warning Messages
Name
1
Message Text Description
Warnings
WSYSRES SYSTEM RESET
Instrument was power-cycled or the CPU
was reset.
WDATAINIT DATA INITIALIZED Data storage was erased.
WCONFIGINIT CONFIG INITIALIZED
Configuration storage was reset to factory
configuration or erased.
WO2ALARM1
3
O2 ALARM 1 WARN O
2
concentration alarm limit #1 exceeded
WO2ALARM2
3
O2 ALARM 2 WARN O
2
concentration alarm limit #2 exceeded
WCO2ALARM1
2
CO2 ALARM 1 WARN
CO
2
concentration alarm limit #1
exceeded
WCO2ALARM2
2
CO2 ALARM 2 WARN
CO
2
concentration alarm limit #2
exceeded
WSAMPFLOW SAMPLE FLOW WARN
Sample flow outside of warning limits
specified by SAMP_FLOW_SET variable.
WSAMPPRESS SAMPLE PRESS WARN
Sample pressure outside of warning limits
specified by SAMP_PRESS_SET
variable.
WBOXTEMP BOX TEMP WARNING
Chassis temperature outside of warning
limits specified by BOX_SET variable.
WO2CELLTEMP
3
O2 CELL TEMP WARN
O
2
sensor cell temperature outside of
warning limits specified by O2_CELL_SET
variable.
WCO2CELLTEMP
2
CO2 CELL TEMP WARN
CO
2
sensor cell temperature outside of
warning limits specified by
CO2_CELL_SET variable.
WDYNZERO CANNOT DYN ZERO
Contact closure zero calibration failed
while DYN_ZERO was set to ON.
WDYNSPAN CANNOT DYN SPAN
Contact closure span calibration failed
while DYN_SPAN was set to ON.
WREARBOARD REAR BOARD NOT DET
Rear board was not detected during
power up.
WRELAYBOARD RELAY BOARD WARN
Firmware is unable to communicate with
the relay board.
WFRONTPANEL FRONT PANEL WARN
Firmware is unable to communicate with
the front panel.
WANALOGCAL ANALOG CAL WARNING
The A/D or at least one D/A channel has
not been calibrated.
1
The name is used to request a message via the RS-232 interface, as in “T BOXTEMP”.
2
T-Series/E-Series: 801, 803 or 802 with CO
2
option.
3
T-Series/E-Series: 802 or 803.
4
T-Series/E-Series: 801 or 803.
5
T-Series/E-Series: 802 only.
6
T-Series/E-Series: 803 only.
7
T-Series/E-Series: 802 with CO
2
option.
8
T-Series/E-Series: 801 or 802 without CO
2
option.
10
External analog input option.
07275B DCN6418