A SERVICE OF

logo

Teledyne API Model T360/T360M Operation Manual Getting Started
59
2. Set the expected CO
2
span gas concentration
This sequence causes the
analyzer to prompt for the
expected CO
2
span
concentration.
M-P CAL CO2 SPAN CONC: 400.000 Conc
0 0 0 4 5 .0 ENTR EXIT
M-P CAL RANGE = 500.000 PPM CO2 =X.XXX
< TST TST > ZERO CONC EXIT
SAMPLE* RANGE = 500.000 PPM CO2 =X.XXX
< TST TST > CAL SETUP
The CO
2
span
concentration values
automatically default to
400.0 Conc.
To change this value to
the actual concentration of
the span gas, enter the
number by pressing the
button under each digit
until the expected value
appears.
EXI
T
ignores the new setting
and returns to the previous
display.
ENTR accepts the new setting
and returns to the
previous display..
NOTE
For this Initial Calibration it is important to independently verify the precise CO
2
Concentration
Value of the SPAN gas.
If the source of the Span Gas is from a Calibrated Bottle, use the exact concentration value
printed on the bottle.
07272B DCN6552