![](https://pdfstore-manualsonline.prod.a.ki/pdfasset/0/71/07113e5e-2910-45dc-b658-c59ee6d49225/07113e5e-2910-45dc-b658-c59ee6d49225-bgb5.png)
Teledyne API Model T360/T360M Operation Manual Calibration Procedures
179
5.5.1. Zero/Span Calibration on Auto Range or Dual Ranges
If the analyzer is being operated in dual range mode or auto range mode, then the
high and low ranges must be independently calibrated.
When the analyzer is in either dual or auto range modes the user must run a
separate calibration procedure for each range. After pressing the CAL, CALZ or
CALS buttons the user is prompted for the range that is to be calibrated as seen in
the CALZ example below:
M-P CAL STABIL=XXX.X PPM CO2 =XXX.X
< TST TST > CAL CALZ CALS SETUP
ACT ION:
Allow zero gas to enter the sample port at the
rear of the instrument.
Wait until ST ABIL
falls below 1.0 ppm.
This may take several
minutes.
SAMPLE* RANGE = 500.000 PPM CO2 =XXX.X
<TST TST> CAL CALZ CALS SETUP
SAMPLE STABIL=XXX.X PPM CO2 =XXX.X
< TST TST > CAL CALZ CALS SETUP
Set the Display to show the
STABIL test function.
This function calculates the
stability of the CO
2
measurement
Continue Calibration as per
Standard Procedure
ZERO CAL M RANGE = 500.000 PPM CO2 =XXX.X
< TST TST > ZERO SPAN CONC EXIT
ANALYZER ENTERS
ZERO CAL MODE
SAMPLE RANGE TO CAL: LOW
LOW HIGH ENTR SETUP
SAMPLE RANGE TO CAL: HIGH
LOW HIGH ENTR SETUP
Once this selection is made, the calibration procedure continues as previously
described in Section 5.2. The other range may be calibrated by starting over from
the main SAMPLE display.