Teledyne API - T360/T360M, 360E/360EM Appendix A Menu Trees (05233C DCN6552)
A-40
MODBUS
Register Address
(dec., 0-based)
Description Units
103
1,2
O
2
concentration limit 2 exceeded
200
3
In CO
2
calibration mode
201
1,3
CO
2
concentration limit 1 exceeded
202
1,3
CO
2
concentration limit 2 exceeded
MODBUS Coil Registers
(single-bit; read/write)
0 Maps to relay output signal 36 (MB_RELAY_36 in signal I/O list)
1 Maps to relay output signal 37 (MB_RELAY_37 in signal I/O list)
2 Maps to relay output signal 38 (MB_RELAY_38 in signal I/O list)
3 Maps to relay output signal 39 (MB_RELAY_39 in signal I/O list)
20
4
Triggers zero calibration of range #1 (on enters cal.; off exits cal.)
21
4
Triggers span calibration of range #1 (on enters cal.; off exits cal.)
22
4
Triggers zero calibration of range #2 (on enters cal.; off exits cal.)
23
4
Triggers span calibration of range #2 (on enters cal.; off exits cal.)
1
Optional
2
O
2
option
3
CO
2
option
4
Set DYN_ZERO or DYN_SPAN variables to ON to enable calculating new slope or offset. Otherwise a calibration check
is performed.
10
External analog input option.
11
Except T300U, 300EU2 and M320EU configurations.