Troubleshooting and Service Teledyne API Model T360/T360M Operation Manual
256
8.6. REPAIR PROCEDURES
This contains procedures that might need to be performed on rare occasions when
a major component of the analyzer requires repair or replacement.
8.6.1. Repairing Sample Flow Control Assembly
The critical flow orifice is housed in the flow control assembly (Teledyne API
part number: 001760400) located on the top of the optical bench. A sintered filter
protects the jewel orifice so it is unusual for the orifice to need replacing, but if it
does, or the filter needs replacement please use the following procedure (see the
Spare Parts list in Appendix B for part numbers and kits):
1. Turn off power to the analyzer.
2. Locate the assembly attached to the sample pump, see Figure 3-5.
3. Disco
nnect the pneumatic connection from the flow assembly and the
assembly from the pump.
4. Remove the fitting and the components as shown in the exploded view in
Figure 8-10.
5. Repl
ace the o-rings (p/n OR000001) and the sintered filter (p/n FL000001).
6. If replacing the critical flow orifice itself (P/N 00094100), make sure that the
side with the colored window (usually red) is facing upstream to the flow gas
flow.
7. Apply new Teflon
®
tape to the male connector threads
8. Re-assemble in reverse order.
9. After reconnecting the power and pneumatic lines, flow check the instrument
as described in Section 6.3.3.