Teledyne API T703/T703U Calibrator Operation Manual General Troubleshooting & Service
187
If the above voltages check out, it is more likely that a problem is due to the UV Lamp
than due to the Lamp Power Supply. Replace the UV Lamp and if the problem persists,
replace the Lamp Power Supply.
8.6.2. TROUBLESHOOTING THE O
3
GENERATOR IN THE T703U
(See Section 8.6.1 for troubleshooting the T703 O
3
generator).
1. Make sure the calibrator is idling in STANDBY mode.
2. Unplug the cable connector at P1 on the Lamp Power Supply and confirm that
+15VDC is present between Pins 1 and 2 on the cable connector.
3. If this voltage is incorrect, check the DC test points on the relay PCA as described
in Section 8.4.3.
4. Remove the
cover of the photometer (see Figure 9-23) and check for the presence
of the followin
g
voltages on the UV lamp power supply PCA:
+800 mVDC ± 10 mVDC between TP1 and TP4 (grnd)
If this voltage is incorrect, either the UV lamp power supply PCA is faulty or the
I
2
C bus is not communicating with the UV lamp power supply PCA.
+5VDC between TP3 and TP4 (grnd)
If this voltages is less than 4.8 or greater than 5.25, either the 5 VDC power
supply or the UV lamp power supply PCA are faulty.
If the above voltages check out, it is more likely that a problem is due to the UV Lamp
than due to the Lamp Power Supply. Replace the UV Lamp and if the problem persists,
replace the Lamp Power Supply.
8.7. REPAIR PROCEDURES
8.7.1. REPAIRING SAMPLE FLOW CONTROL ASSEMBLY
The critical flow orifice is housed in the flow control assembly (Teledyne API part
number: 001760400) located on the top of the optical bench. A sintered filter protects
the jewel orifice so it is unusual for the orifice to need replacing, but if it does, or the
filter needs replacement please use the following procedure (see the Spare Parts list in
Appendix B for part numbers and kits):
1. Turn off power to the calibrator.
2. Locate the assembly to be repaired, see Figure 9-1 (T703) or Figure 9-2 (T703U).
3.
Disco
nnect the pneumatic connection from the flow assembly.
4. Remove the fitting and the components as shown in the exploded view in Figure
8-5.
5.
Repl
ace the o-rings (p/n OR000001) and the sintered filter (p/n FL000001).
6. If replacing the critical flow orifice itself (P/N 000941000), make sure that the side
with the colored window (usually red) is facing downstream to the gas flow.
7. Apply new Teflon
®
tape to the male connector threads.
8. Reassemble in reverse order.