A SERVICE OF

logo

Communications Teledyne API T703/T703U Calibrator Operation Manual
114
Figure 5-3: Jumper and Cables for Multidrop Mode
4. Close the instrument.
5. Referring to Figure 5-4, use straight-through DB9 male-DB9 female cables to
interconnect the host RS
232 port to the first instrument’s RS232 port; then from the
first instrument’s COM2 port to the second instrument’s RS232 port; from the
second instrument’s COM2 port to the third instrument’s RS232 port, etc.,
connecting in this fashion up to eight instruments, subject to the distance limitations
of the RS-232 standard.
6. On the rear panel of each instrument, adjust the DCE DTE switch so that the green
and the red LEDs (RX and TX) of the COM1 connector (labeled RS232) are both lit.
(Ensure you are using the correct RS-232 cables internally wired specifically for RS-
232 communication).
7. BEFORE communicating from the host, power on the instruments and check that
the Machine ID (Section 5.6.2) is unique for each.
a.
In the SETUP Model me
nu, use SETUP>MORE>COMM>ID. The default ID is
typically either the model number or “0”.
b. To change the identification number, press the button below the digit to be changed.
c. Press/select ENTR to accept the new ID for that instrument.
8. Next, in the SETUP>MORE>COMM>COM1 menu (do not use the COM2 menu for
multidrop), edit the COM1 MODE parameter as follows: press/select EDIT and set
only QUIET MODE, COMPUTER MODE, and MULTIDROP MODE to ON. Do not
change any other settings.
9. Press/select ENTER to accept the changed settings, and ensure that COM1 MODE
now shows 35.
07223C DCN6572