Teledyne API T703/T703U Calibrator Operation Manual Maintenance Schedule & Procedures
157
6. Replace the cover on the instrument.
Figure 7-3: Photometer – Location of UV Detector Gain Adjustment & UV Lamp Set Screw
7.6. PHOTOMETER UV SOURCE LAMP REPLACEMENT
This procedure details the steps for replacement of the UV source lamp in the optical
bench assembly. This procedure should be done whenever the lamp can no longer be
adjusted as described in Section 10.2.3.
1. Turn the instrument off.
2. Remove the cover from the instrument.
3. Locate the Optical Bench Assembly (see Figure 3-5)
4. Locate the UV lamp at the rear of
the optical bench assembly (see Figure 7-3)
5.
Unpl
ug the lamp cable from the power supply connector on the side of the optical
bench.
6. Slightly loosen (do not remove) the UV lamp setscrew and pull the lamp from its
housing.
NOTE
The UV lamp contains mercury (Hg), which is considered hazardous waste. The lamp should be
disposed of in accordance with local regulations regarding waste containing mercury.
7. Install the new lamp in the housing, pushing it all the way in. Leave the UV lamp
setscrew loose for now.
8. Turn the instrument back on and allow it to warm up for at least 15 minutes.
9. Turn the UV detector gain adjustment pot (See Figure 7-3) clockwise to its minimum
value. The pot shoul
d cli
ck softly when the limit is reached.
10. Perform the UV Lamp Adjustment procedure described in Section 7.5, with the
following exceptions:
11. Slowly rotate
the lamp in its housing (up to ¼ turn in either direction) until a
MAXIMUM
value (or 4600 mVDC) is observed.
Make sure the lamp is pushed all the way into the housing while performing this
rotation.
If the PHOTO_DET will not drop below 5000 mV while performing this rotation,
contact T-API Technical Support for assistance.