A SERVICE OF

logo

General Troubleshooting & Service Teledyne API T703/T703U Calibrator Operation Manual
178
8.4.9.2. Test Channel / Analog Outputs Voltage
To verify that the analog output is working properly, connect a voltmeter to the output in
question and perform an analog output step test as follows:
For each of the steps the output should be within 1% of the nominal value listed in the
table below except for the 0% step, which should be within 0mV ±2 to 3 mV. Make
sure you take into account any offset that may have been programmed into channel (See
Section 4.9.1.5).
07223C DCN6572