A SERVICE OF

logo

Operating Instructions Teledyne API – T101 Operation Manual
86
The following DC current output limits apply to the current loop modules:
Table 4-11. Analog Output Current Loop Range
RANGE MINIMUM OUTPUT MAXIMUM OUTPUT
0-20 mA 0 mA 20 mA
These are the physical limits of the current loop modules, typical applications use 2-20 or 4-20
mA for the lower and upper limits. Please specify desired range when ordering this option.
The default offset for all ranges is 0 mA.
Pin assignments for the output connector at the rear panel of the instrument are shown in
Table 4-12.
ANALOG OUT
A1 A2 A3 A4
+ - + - + - + -
Table 4-12. Analog Output Pin Assignments
PIN ANALOG
OUTPUT
VOLTAGE
SIGNAL
CURRENT
SIGNAL
1 V Out I Out +
2
A1
Ground I Out -
3 V Out I Out +
4
A2
Ground I Out -
5 A3 Not Used Not Used
7 V Out not available
8
A4
Ground not available
See Figure 3-4 for a the location of the analog output connector on the instrument’s rear
panel.
07266B DCN6485