A SERVICE OF

logo

Teledyne API – T101 Operation Manual Operating Instructions
99
4.6.6. LAMP CALIBRATION
An important factor in accurately determining H
2
S concentration, once the H
2
S is
converted to SO
2,
is the amount of UV light available to transform the SO
2
into SO
2
* (see
Sections 10.1.1 and 10.1.2). The Model T101 compensates for variations in the intensity
of the available UV light by
adjusting the H
2
S concentration calculation using a ratio
(LAMP RATIO)that results from dividing the current UV lamp (UV LAMP) intensity
by a value stored in the CPU’s memory (LAMP_CAL). Both LAMP Ratio and UV
Lamp are test functions viewable from the instrument’s front panel.
To cause the analyzer to measure and record a value for LAMP_CAL, press:
Exit at
any time
to return
to main
the
SETUP
menu
DIAG SIGNAL I / O
NEXT ENTR EXIT
Repeat Pressing NEXT until . . .
DIAG LAMP CALIBRATION
PREV NEXT ENTR EXIT
DIAG FCAL LAMP CAL VALUE:4262.4 mV
4 2 6 2 .4 ENTR EXIT
ENTR accepts the
new value
EXIT ignores the new
value
The value displayed is the
current output of the UV
source reference detector
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL SETUP
SAMPLE ENTER SETUP PASS : 818
8 1 8 ENTR EXIT
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
07266B DCN6485