A SERVICE OF

logo

Troubleshooting & Service Model T101 Instruction Manual
218
Toggles parameter
ON/OFF
Exit returns to
DIAG display & all values
return to software control
DIAG SIGNAL I/O
PREV NEXT ENTR EXIT
DIAG I/O 0 ) EXT_ZERO_CAL=ON
PREV NEXT JUMP PRNT EXIT
DIAG I/O 23) ST H2S_MODE=ON
PREV NEXT JUMP ON PRNT EXIT
DIAG I/O 37) SAMPLE_PRESSURE=6000.0 MV
PREV NEXT JUMP PRNT EXIT
DIAG I/O 23) ST H2S_MODE OFF
PREV NEXT JUMP OFF PRNT EXIT
If parameter is an
input signal
If parameter is an output
signal or control
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
SAMPLE ENTER SETUP PASS : 818
8 1 8 ENTR EXIT
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SAMPLE RANGE = 500.0 PPB H2S =XXX.X
< TST TST > CAL SETUP
Figure 9-2. Example of Signal I/O Function
9.1.4. STATUS LEDS
Several color-coded, light-emitting diodes (LEDs) are located inside the instrument to
determine if the analyzer’s CPU, I
2
C communications bus and relay board are
functioning properly.
07266B DCN6485