Calibration Procedures Teledyne API – T101 Operation Manual
170
6.4. MANUAL CALIBRATION WITH ZERO/SPAN VALVES
Zero and Span calibrations using the Zero/Span Valve option are similar to that described
in Section 6.2 except that:
Zero air and span gas is supplied
to the analyzer through the zero gas and span gas inlets
rather than through the sample inlet.
The zero and cal operations are initiated directly and independently with dedicated
buttons (CALZ & CALS)
STEP ONE: As shown below connect the sources of zero air and span gas to the
respective ports on the rear panel (Figure 3-2).
External Zero
Air Scrubber
Filter
MODEL 701
Zero Air
Generator
Source of
SAMPLE Gas
VENT if input is pressurized
MODEL T700
Gas Dilution Calibrator
Chassis
Sam
p
le
Zero Air
Exhaust
Span
Calibrated
H
2
S gas
(At high
concentration)
Figure 6-2. Setup for Manual Calibration with Z/S Valve Option Installed
07266B DCN6485