A SERVICE OF

logo

Getting Started Teledyne API T801 NDIR CO2 Analyzer Operation Manual
32
3.3. CONNECTIONS AND SETUP
This section presents the electrical (Section 3.3.1) and pneumatic (Section 3.3.2)
connections for setup and preparing for instrument operation.
3.3.1. ELECTRICAL CONNECTIONS
Note To maintain compliance with EMC standards, it is required that the
cable length be no greater than 3 meters for all I/O connections,
which include Analog In, Analog Out, Status Out, Control In,
Ethernet/LAN, USB, RS-232, and RS-485.
This section presents the electrical connections for AC power and
communications.
3.3.1.1. CONNECTING POWER
WARNING - ELECTRICAL SHOCK HAZARD
High Voltages are present inside the analyzers case.
Turn OFF analyzer power before disconnecting or connecting PCAs, wiring
harnesses or electrical subassemblies.
Power connection must have functioning ground connection.
Do not defeat the ground wire on power plug.
Do not operate with cover off.
Attach the power cord to the analyzer and plug it into a power outlet capable of
carrying at least 10 A current at your AC voltage and that it is equipped with a
functioning earth ground.
CAUTION
GENERAL SAFETY HAZARD
The T801 analyzer can be configured for both 100-130 V and 210-240 V at either
47 or 63 Hz.
To avoid damage to your analyzer, make sure that the AC power voltage
matches the voltage indicated on the Analyzer’s serial number label (See
Figure 3-4) before plugging the T801 into line power.
07274B DCN6418