12.5.9. CALIBRATION - SLOPE AND OFFSET
Calibration of the analyzer is performed exclusively in software.
During instrument calibration the user enters expected values for span via the
front panel keypad; values are not entered during a zero operation, commanding
the instrument to make readings of calibrated sample gases for both levels. The
readings taken are adjusted, linearized, and compared to the expected values.
With this information the software computes values for instrument slope and
offset and stores these values in memory for use in calculating concentrations of
the sample gas.
The instrument slope and offset values recorded during the last calibration are
available for viewing from the from the front panel (See Section 3.4.2)
12.5.10. TEMPERATURE AND PRESSURE COMPENSATION
Changes in ambient pressure can have a noticeable effect on the CO
2
concentration calculations. To account for this, the T801 software includes a
feature that allows the instrument to compensate CO
2
calculations based on
changes in ambient pressure. The CO
2
sensor is housed inside a temperature
controlled manifold. This minimizes temperature effects on the measured
concentrations.
12.5.11. INTERNAL DATA ACQUISITION SYSTEM (DAS)
The DAS is designed to implement predictive diagnostics that stores trending
data for users to anticipate when an instrument will require service. Large
amounts of data can be stored in non-volatile memory and retrieved in plain text
format for further processing with common data analysis programs. The DAS
has a consistent user interface in all Teledyne API analyzers. New data
parameters and triggering events can be added to the instrument as needed.
Depending on the sampling frequency and the number of data parameters the
DAS can store several months of data, which are retained even when the
instrument is powered off or a new firmware is installed. The DAS permits users
to access the data through the instrument’s front panel or the remote interface.
The latter can automatically download stored data for further processing. For