![](https://pdfstore-manualsonline.prod.a.ki/pdfasset/9/c7/9c770034-caf1-4867-8b1e-24bde634536b/9c770034-caf1-4867-8b1e-24bde634536b-bg155.png)
T700, M700E Calibrator Manuals APPENDIX A-3: Warnings and Test Functions, Revision D.3 (05623D DCN5839)
A-23
Table A-3: Test Functions, Revision D.3
Name
1
Message Text Description
Test Measurements
Parmeter name in
T-Series
and in E-Series w/
software v. D.3
and higher.
Parameter name in earlier E-Series software
versions and test meas value
A-CAL ACT CAL=0.0800 LPM Actual cal. gas flow rate.
T-CAL TARG CAL=0.0000 LPM Target cal. gas flow rate.
A-DIL ACT DIL=1.920 LPM Actual diluent flow rate.
T-DIL TARG DIL=0.000 LPM Target diluent flow rate.
O3GENREF
6
O3 GEN REF=1000.0 MV O
3
generator reference detector reading.
O3FLOW
6
O3 FLOW=0.1050 LPM
O
3
generator flow rate. Note: this is simply a constant,
specified by the O3_GEN_FLOW variable.
O3GENDRV
6
O3 GEN DRIVE=800.0 MV O
3
generator lamp drive output.
O3LAMPTMP
6
O3 LAMP TEMP=49.7 C O
3
generator lamp temperature.
CAL PRES CAL PRESSURE=25.1 PSIG Cal. gas pressure.
DIL PRES DIL PRESSURE=25.1 PSIG Diluent pressure.
REG PRES REG PRESSURE=20.1 PSIG Regulator pressure.
(n/a) ACT=GENERATE 37 PPB O3
Actual concentration being generated, computed from
real-time inputs.
A-GAS (n/a) T-GAS ± 1%
T-GAS (n/a)
A-O3
7
(n/a) T-O3 ± 1%
T-O3
7
(n/a)
(n/a) TARG=GENERATE 100 PPB O3 Target concentration to generate.
T-FLW (n/a) Target standard flow in LPM
BOX TMP BOX TEMP=31.2 C Internal chassis temperature.
PERM1 TMP PERM TUBE #1 TEMP=50.4 C Permeation tube #1 temperature.
PERM2 TMP
3
PERM TUBE #2 TEMP=50.4 C Permeation tube #2 temperature.
PERMFLW
3
PERM FLOW=0.1050 LPM
Permeation tube flow rate. This is a property of the
permeation tube (SETUP-GAS-PERM). Its value
depends on which permeation tube is in use.
PH MEAS
7
PHOTO MEASURE=2998.8 MV Photometer detector measure reading.
PH REF
7
PHOTO REFERENCE=3000.0 MV Photometer detector reference reading.
PH FLW
7
PHOTO FLOW=0.2978 LPM Photometer sample flow rate.
PH LTEMP
7
PHOTO LAMP TEMP=52.6 C Photometer lamp temperature.
PH PRES
7
PHOTO SPRESS=29.9 IN-HG-A Photometer sample pressure.
PHO STEMP
7
PHOTO STEMP=31.8 C Photometer sample temperature.
PHO SLOPE
7
PHOTO SLOPE=1.000
Photometer slope computed during zero/span bench
calibration.
PH OFFST
7
PHOTO OFFSET=0.0 PPB
Photometer offset computed during zero/span bench
calibration.
PHOTOSTABIL
2
PHOTO STABIL=0.1 PPB
Photometer concentration stability (standard deviation of
25 bench concentration samples taken 10 seconds
apart).