![](https://pdfstore-manualsonline.prod.a.ki/pdfasset/9/c7/9c770034-caf1-4867-8b1e-24bde634536b/9c770034-caf1-4867-8b1e-24bde634536b-bga3.png)
Teledyne API – Model T700 Dynamic Dilution Calibrator Overview of Operating Modes and Basic Operation
161
AUTOMATIC TEST CHANNEL CALIBRATION
Note
Before performing this procedure, ensure that the AUTO CAL feature is turned
OFF for MFC_DRIVE_1, MFC_DRIVE_2 and MFC_DRIVE_3 if installed). Ensure
that the AUTO CAL feature is turned ON for the TEST CHANNEL (see Section
4.10.1.6).
To calibrate the outputs as a group with the AOUTS CALIBRATION command, press:
From the
(See Section 6.9.1.1.)
Analyzer
automatically
calibrates all
channels for which
is turned
If any of the channels have not
been calibrated ot if at least one
channel has AUTO-CAL turned
OFF, this message will read
.
DIAG AIO NOT AUTO CAL. MFC_DRIVE_1
DIAG AIO NOT AUTO CAL. MFC_DRIVE_2
This message
appears when
is
Turned
for
a channel
EXIT
DIAG
PREV NEXT EXIT
DIAG AIO NOT AUTO CAL. MFC_DRIVE_3
Note
Manual calibration should be used for the 0.1V range or in cases where the
outputs must be closely matched to the characteristics of the recording device.
Manual calibration requires that the AUTOCAL feature be disabled.