![](https://pdfstore-manualsonline.prod.a.ki/pdfasset/9/c7/9c770034-caf1-4867-8b1e-24bde634536b/9c770034-caf1-4867-8b1e-24bde634536b-bg149.png)
T700, M700E Calibrator Manuals APPENDIX A-1: Software Menu Trees, Revision D.3 (05623D DCN5839)
A-11
1
Only occurs if one of the voltage ranges is selected.
2
Manual adjustment menu only appears if either the AUTO CAL feature is OFF for the selected output or the RANGE is set for CURRent.
3
Only appears if optional O
3
photometer bench is installed.
4
Only appears if optional O
3
generator is installed
5
Only appears if optional 2nd Cal Gas MFC is installed
6
DIAG Menu is inactive while instrument is in GENERATE mode.
7
DO NOT Edit the settings for the MFC drives!
MORE DIAG
6
ENTER PASSWORD: 818
SIGNAL I/O
ANALOG
OUTPUT
ANALOG I/O
CONFIGURATION
PHOTO FLOW
SENSOR CAL
3
MFC
CONFIGURATION
O3 GEN
4
CALIBRATION
PRESSURE
CALIBRATION
AUTO LEAK
CHECK
TEST CHAN
OUTPUT
Press ENTR
to start test
NEXTPREV
0)- 11) CONTROL_IN_1
THRU 12
12)-23) CONTROL_OUT_1
THRU 12
54 INTERNAL ANALOG
to VOLTAGE SIGNALS
72 (see Appendix A)
ON
OFF
24) ST_SYSTEM_OK
25) ST_CAL_ACTIVE
26) ST_DIAG_MODE
27) ST_TEMP_ALARM
28) ST_PRESS_ALARM
29) PERM_VALVE_1
30) PERM_HTR_1
31) VENT_VALVE
32) RELAY_WATCHDOG
33) GPT_VALVE
34) PHOTO_REF_VALVE
35) O3_GEN_VALVE
36) O3_PUMP_ON
37) OUTPUT_VALVE_A
38) OUTPUT_VALVE_B
39) PHOTO_LAMP_HEATER
40) O3_GEN_HEATER
41) VALVE_WATCHDOG
42) CYL_VALVE_1
43) CYL_VALVE_2
44) CYL_VALVE_3
45) CYL_VALVE_4
46) PURGE_VALVE
47) INPUT_VALVE
48) MAINT_MODE
49) LANG2_SELECT
50) SEQUENCE_LED
51) AUTO_TIMER_LED
52) FAULT_LED
53) AUDIBLE_BEEPER
AOUTS CALIBRATED AIN CALIBRATED
SET>
<SET
CAL
Initiates auto-calibration of all
analog outputs
Initiates auto-calibration of the analog inputs’
zero and span points
CAL
EDIT
RANGE
10V
5V1V0.1V
ON
OFF
OVER
RANGE
REC
OFFSET
1
AUTO
1
CAL
ON
OFF
CAL
CALIBRATED
Sets a voltage
offset for the
output
Auto Cal
1
U100 UP10 UP DOWN DN10 D100
Manual Cal
2
Initiates auto-calibration
of the selected analog input
Initiates Internal leak check
(See Chapter 11 for more
information)
NONE
O3 PHOTO MEAS
O3 PHOTO REF
O3 GEN REF
SAMPLE PRESSURE
SAMPLE FLOW
SAMPLE TEMP
PHOTO LAMP TEMP
O3 LAMP TEMP
CHASSIS TEMP
O3 PHOTO CONC
See SECONDARY SETUP Menu
DIAG - CALIBRATION Submenu
PREV NEXT
PREV NEXT
MAIN MENU
SETUP
Backpressure
Compensation
Figure A-8: SECONDARY SETUP Menu; DIAG Submenu – GAS CONFIGURATION