![](https://pdfstore-manualsonline.prod.a.ki/pdfasset/1/43/143577a8-6851-4124-b1df-dcd55bb6005b/143577a8-6851-4124-b1df-dcd55bb6005b-bg31.png)
Teledyne API – Model T300/T300M CO Analyzer Getting Started
47
3.3.1.2. CONNECTING ANALOG INPUTS (OPTION)
The Analog In connector is used for connecting external voltage signals from other
instrumentation (such as meteorological instruments) and for logging these signals in the
analyzer’s internal DAS. The input voltage range for each analog input is 0-10 VDC.
Figure 3-8: Analog In Connector
Pin assignments for the Analog In connector are presented in Table 3-4.
Table 3-4: Analog Input Pin Assignments
PIN DESCRIPTION
DAS
PARAMETER
1
1 Analog input # 1 AIN 1
2 Analog input # 2 AIN 2
3 Analog input # 3 AIN 3
4 Analog input # 4 AIN 4
5 Analog input # 5 AIN 5
6 Analog input # 6 AIN 6
7 Analog input # 7 AIN 7
8 Analog input # 8 AIN 8
GND Analog input Ground N/A
1
See Section 7 for details on setting up the DAS.
3.3.1.3. CONNECTING ANALOG OUTPUTS
The T300 is equipped with several analog output channels accessible through a
connector on the back panel of the instrument. The standard configuration for these
outputs is mVDC. An optional current loop output is available for each.
When the instrument is in its default configuration, channels A1 and A2 output a signal
that is proportional to the CO concentration of the sample gas. Either can be used for
connecting the analog output signal to a chart recorder or for interfacing with a
datalogger.
Output A3 is only used on the T300/T300M if the optional CO
2
or O
2
sensor is installed.
Channel A4 is special. It can be set by the user (see Section 5.9.8.1) to output any one
of the parameters a
ccessible through the <TST TST> buttons of the units sample
display.
To access these signals attach a strip chart recorder and/or data-logger to the appropriate
analog output connections on the rear panel of the analyzer.